A Fast, Novel and Analytically Sensitive Clinical Research Method for the Analysis of Free Testosterone using UPLC-MS/MS

Buy Lab Tests Online

madman

Super Moderator


Abstract

Background


The aim of this work was to develop an analytically sensitive, rapid clinical research method for the analysis of free testosterone in serum using a small sample volume. Laboratories have found this difficult, partly due to the inherent low concentrations (less than 3% of total testosterone is free), technical challenges and variability in results have been noted with equilibrium dialysis, which is considered the gold standard.


Methods

Testosterone certified reference material (Merck, UK) was used to create calibrators in 52.75 mM HEPES buffer adjusted to pH 7.4 to mimic the ionic environment of serum. Precision was evaluated using in-house QC materials prepared in serum (Golden West Biologicals, USA and BioIVT, UK). 200 µL serum was incubated at 37°C with 400 µL pH-adjusted 52.75 mM HEPES buffer for 2 hours, mixing at 800 r.p.m. with Rapid Equilibrium Analysis (RED) inserts and plates (Thermo Scientific™, UK). Testosterone-13C3 (Merck, UK) was spiked into resultant diasylate, then liquid-liquid extraction (LLE) was used to concentrate and clean the sample. The ACQUITY™ UPLC™ I-Class System with the Waters™ ACQUITY UPLC BEH C18, 1.7µm, 2.1 x 100 mm Column using a water/methanol/ammonium fluoride gradient and the Waters Xevo™ TQ Absolute Mass Spectrometer were used for quantification. An accelerated protocol for equilibrium dialysis, using a 2-hour incubation and fast mixing step was used (4 hours or longer is more typical historically) though it is essential to control the pH of the dialysis buffer and, particularly, the temperature of the chamber.


Results

No significant carryover or matrix effects were demonstrated with linearity from 0.5-650 pg/mL. Addition of high concentrations of endogenous substances (albumin, bilirubin, creatinine, cholesterol, triglycerides and uric acid) did not affect quantification. Analytical sensitivity investigations indicated the analytical sensitivity would allow precise quantification (≤20%CV and ≤15% bias) at 0.50 pg/mL. Coefficients of variation (CV) for total precision and repeatability on 5 analytical runs for low, mid and high QCs at approximately 2.79, 8.80 and 134 pg/mL respectively were all ≤ 8.4% (n=25). Accuracy was assessed by analyzing 45 male external quality assurance samples (NEQAS, UK) ranging in concentration from 50.1-268.6 pg/mL and the Bland-Altman agreement showed a small negative bias of -7.6% compared with the all-laboratory trimmed mean (ALTM), though excellent agreement overall. When LC-MS/MS determined concentrations were processed against Vermeulen equation determined concentrations (total testosterone and SHBG concentrations were provided by the scheme, a concentration of 45 g/L albumin was assumed), again good agreement was shown by the Bland-Altman bias of -7.7%.


Conclusions

A novel clinical research method using a fast equilibrium dialysis procedure with LLE and LC-MS/MS analysis with only 200 µL of serum has been demonstrated for clinical research use. Sub pg/mL analytical sensitivity, excellent linearity and precision, with minimal matrix effects and a strong agreement with an EQA scheme were established.
 

Attachments

  • hvae106.043.pdf
    91.3 KB · Views: 26
Defy Medical TRT clinic doctor
* Coefficients of variation (CV) for total precision and repeatability on 5 analytical runs for low, mid and high QCs at approximately 2.79, 8.80 and 134 pg/mL respectively were all ≤ 8.4% (n=25). Accuracy was assessed by analyzing 45 male external quality assurance samples (NEQAS, UK) ranging in concentration from 50.1-268.6 pg/mL and the Bland-Altman agreement showed a small negative bias of -7.6% compared with the all-laboratory trimmed mean (ALTM), though excellent agreement overall. When LC-MS/MS determined concentrations were processed against Vermeulen equation determined concentrations (total testosterone and SHBG concentrations were provided by the scheme, a concentration of 45 g/L albumin was assumed), again good agreement was shown by the Bland-Altman bias of -7.7%.
 
Buy Lab Tests Online
Defy Medical TRT clinic

Sponsors

bodybuilder test discounted labs
cheap enclomiphene
TRT in UK Balance my hormones
Discounted Labs
Testosterone Doctor Near Me
Testosterone books nelson vergel
Register on ExcelMale.com
Trimix HCG Offer Excelmale
BUY HCG CIALIS

Online statistics

Members online
1
Guests online
1
Total visitors
2

Latest posts

Top